DEAR
PLEASE COULD YOU GUIDE US HOW TO RUN OUR INSTRUMENT WITH SDVD Mode so anytime we try it but no success and could only select Axial or Radial mode to run the samples and we don't know which configuration should we have to do for that ?
Find maintenance, best practice, and troubleshooting articles.
DEAR
PLEASE COULD YOU GUIDE US HOW TO RUN OUR INSTRUMENT WITH SDVD Mode so anytime we try it but no success and could only select Axial or Radial mode to run the samples and we don't know which configuration should we have to do for that ?
Hello,
Is your instrument a 5800 or a 5900 ICP-OES? SVDV is only available with the 5900 ICP-OES.
It is technically possible to upgrade a 5800 to be able to run in SVDV- but this is something you'll need to contact your local Agilent Atomic spectroscopy team to get a quote for. It is a paid upgrade and will also require an onsite visit from an Agilent Field Service representative
Dear John
thank you for your response and we weren't away that there is a differency between 5800 &5900 so ours is 5800 .
then we realize that when you're using AVS 6 the instrument sensibility's low than when you use without AVS system so can you advise it ?
Hello,
what is the replicate read time on the Conditions page without the AVS?
yes we use Ytterium solution .
however AVS sytem's a bit fast than normal system but detection limit of without AVS is lower and better than with AVS ...?
it's the same 3 times
it's the same 3 times
I'm sorry I don't understand. What is the replicate read time in seconds without the AVS?
Hi Tina
please see for the conditions