ICP-MS 7800 error 1029

Hello, does any know how to resolve the error 1029 water valve sensor error/water valve malfunction. it occurs semi regularly and seems to resolve itself with no noticeable pattern. Restarting the masshunter program, the computer, the chiller, and power cycling are the troubleshooting tools that seem to work eventually. 

Parents
  • Sometimes the valve is just stuck partially open. If you go into your Hardware dashboard, click the sample introduction image, then maintenance. With the chiller on, click the box that says enable temperature control. Click it on and off a couple times, this will cycle the valve open and closed. Then close out of the screen and try to light again. If it doesn’t resolve it the issue persists you will need to log a service ticket and get the valve replaced. 

Reply
  • Sometimes the valve is just stuck partially open. If you go into your Hardware dashboard, click the sample introduction image, then maintenance. With the chiller on, click the box that says enable temperature control. Click it on and off a couple times, this will cycle the valve open and closed. Then close out of the screen and try to light again. If it doesn’t resolve it the issue persists you will need to log a service ticket and get the valve replaced. 