Will a matrix of 2%HNO3 and 0.5%HCL be OK for stability of Ag to prepare 0-300ppb calibration curve or should the HCL rather be 1%?
Find maintenance, best practice, and troubleshooting articles.
Hi All
Will a matrix of 2%HNO3 and 0.5%HCL be OK for stability of Ag to prepare 0-300ppb calibration curve or should the HCL rather be 1%?
If Ag concentration increases, there is the possibility that the HCl can cause AgCl to precipitate (cloudy and white). Risk of nebulizer blockage and/or lower Ag recovery values.