Will a matrix of 2%HNO3 and 0.5%HCL be OK for stability of Ag to prepare 0-300ppb calibration curve or should the HCL rather be 1%?
Hi All
Will a matrix of 2%HNO3 and 0.5%HCL be OK for stability of Ag to prepare 0-300ppb calibration curve or should the HCL rather be 1%?
Hello,
Several application notes are published with only 0.5% HCl added to 1% HNO3 :
Ag in soils / 0.5% HCl + 1% HNO3 : https://www.agilent.com/cs/library/applications/application-soil-7850-icp-ms-5994-2933en-agilent.pdf
Ag in drinking water / 0.5%HCl + 1% HNO3 : https://www.agilent.com/cs/library/applications/5991-4938EN.pdf
Ag in cannabis / 0.5% HCl + 1% HNO3 : https://www.agilent.com/cs/library/applications/5991-8482EN_cannabis_icp-ms_application.pdf
In those application notes, maximum standards of Ag are 100 ppb.
Not 100% sure it wil work the same for 300 ppb, but shoud be OK.
Best regards,
Guillaume
I've used 0.5% up to low ppm levels for Ag and there was enough excess Cl to stabilize it. If you noticed poor recovery with 0.5% that would be your indicator to increase to 1% but I doubt it will be needed.
Hi Guillaume
Thank you for your feedback and references.
Regards.
Hi Cheltajc
Thank you for your answer. Always good to hear from people with experience.
Rergards.
If Ag concentration increases, there is the possibility that the HCl can cause AgCl to precipitate (cloudy and white). Risk of nebulizer blockage and/or lower Ag recovery values.