Hi,
I can see my CPS and CPS RSD, however, when switching to concentration all I get is <0.000 and N/A for my RSD How can I get my raw concentration and RSD?
Thanks for the help!
JJ
Find maintenance, best practice, and troubleshooting articles.
Hi,
I can see my CPS and CPS RSD, however, when switching to concentration all I get is <0.000 and N/A for my RSD How can I get my raw concentration and RSD?
Thanks for the help!
JJ
Hi,
it looks like your sample has a lower concentration then your lowest standard (blank)
if you want to have „negative“ values, you have to remove the „0“ for the minimum concentration in the calibration tab
best Regards
chris
Hi,
it looks like your sample has a lower concentration then your lowest standard (blank)
if you want to have „negative“ values, you have to remove the „0“ for the minimum concentration in the calibration tab
best Regards
chris
the "Min Conc." in the "Calibration" tab should be set to "<None>" for negative values. al other options in the "Min Conc" could give <x.xxx values.