Currently we have our 8900 User Tune set to Custom Tune, and each batch has it's own Tune. Is it better to have User Tune set to Auto Tune and it set to Global Tune?
Find maintenance, best practice, and troubleshooting articles.
Currently we have our 8900 User Tune set to Custom Tune, and each batch has it's own Tune. Is it better to have User Tune set to Auto Tune and it set to Global Tune?
Hi Craig,
Global Tune means, that this tune is used for every batch. If you run for example 5 batches, which are exactly the same, but belongs to 5 different customers, it makes sense to use a global tune setting. Then you only have to tune once. If you have different tune modes in those batches (like one has no HEHe e.g.) it is better not to use the Global Tune function.
I
Hi Craig,
Global Tune means, that this tune is used for every batch. If you run for example 5 batches, which are exactly the same, but belongs to 5 different customers, it makes sense to use a global tune setting. Then you only have to tune once. If you have different tune modes in those batches (like one has no HEHe e.g.) it is better not to use the Global Tune function.