Hello
"2607 Warning: Y-Axis movement disabled due to partially closed top cover" is happening in ICP-MS 7850. It says that try it again after opening the small cover fully. Can anyone show me how to fully open a small cover or solve this problem?
Find maintenance, best practice, and troubleshooting articles.
Hello
"2607 Warning: Y-Axis movement disabled due to partially closed top cover" is happening in ICP-MS 7850. It says that try it again after opening the small cover fully. Can anyone show me how to fully open a small cover or solve this problem?
Hi,
to set the torch box into the maintenance position, the small cover (black) hast to be open.
normally it is enough to open the small cover 90 degree. The instruments detects this with a hall sensor (magnet). You can see this an the left hand side (highlighted)
Sometimes the alignment is a bit off, then you should open the cover until the magnet is triggering the hall sensor. This should solve your problem.
Best Regards
Chris
Hi,
to set the torch box into the maintenance position, the small cover (black) hast to be open.
normally it is enough to open the small cover 90 degree. The instruments detects this with a hall sensor (magnet). You can see this an the left hand side (highlighted)
Sometimes the alignment is a bit off, then you should open the cover until the magnet is triggering the hall sensor. This should solve your problem.
Best Regards
Chris
Thanks for your kind reply. This was helpful~