Dear,
There is a pop up windows saying "F F16505: Error. Instrument (ICB) Offline" while performing the Calibrate Detector (Dark Current Scan). I would be appreciated if anyone here could advise me how to solve this matter. Thank you.
William
Find maintenance, best practice, and troubleshooting articles.
Dear,
There is a pop up windows saying "F F16505: Error. Instrument (ICB) Offline" while performing the Calibrate Detector (Dark Current Scan). I would be appreciated if anyone here could advise me how to solve this matter. Thank you.
William
Hi,
Redo the detector calibration after power cycle the icp-oes and pc. If the same problem is there then contact you local Agilent organization.
however, it's not needed to re-do the detector calibration every time. Only needed during yearly maintenance or after a big repair of the spectrometer.
no problem to work with a former detector calibration/ dark current calibration.