I need to measure 45.5 mass I dont know how to add it to the batch. I am using the mass hunter 4.4V C.01.04
Find maintenance, best practice, and troubleshooting articles.
I need to measure 45.5 mass I dont know how to add it to the batch. I am using the mass hunter 4.4V C.01.04
Hi,
For this function it's needed to activated the additional "Advanced Acquisition" package of the MassHunter. If it's activated then it's possible to select the "Half Mass Acquisition" in "Settings" => "Options".
if the function is selected then it's possible to select +0.5 mass in the "Acq parameters" tab of the batch.
Here some detailed information about the Half Mass setup±
Simplifying Correction of Doubly Charged Ion Interferences with (agilent.com)
if the function isn't available then contact the local Agilent department for more information.
good luck with the setup,
cheers
Edgar