Sample list locked

Hello ALL:

We have  an Agilent ICP-MS 7700X., MassHunter running on windows -7.

My sample list is locked. It is not allow me to  type anything on my sample list.

How Do I unlock my sample list, so that I can modify my sample list.

Your answer will be highly appreciated.

Shah Zaman

Parents
  • Hi Shah, you will need to select the row you want to begin editing and then select “Start Editing Mode”. This is located at the top of the sample list. When selected you have the ability to add and remove anything after the selected row. Once you make the desired changes, “End Editing Mode” otherwise the sample list will not proceed. 

Reply
  • Hi Shah, you will need to select the row you want to begin editing and then select “Start Editing Mode”. This is located at the top of the sample list. When selected you have the ability to add and remove anything after the selected row. Once you make the desired changes, “End Editing Mode” otherwise the sample list will not proceed. 

Children
No Data