I want to run dead time correction on 8900 using 166/164 Er tuning solution. What concentrations should I use?
I want to run dead time correction on 8900 using 166/164 Er tuning solution. What concentrations should I use?
You will need a high and low concentration. The typical dead time standards are 50ppm and 1ppm Er. If you get an error during the dead time adjustment the signal is too high or too low you may need to adjust these concentrations.
You will need a high and low concentration. The typical dead time standards are 50ppm and 1ppm Er. If you get an error during the dead time adjustment the signal is too high or too low you may need to adjust these concentrations.
Hi Cheltajc,
Thanks for your reply. I will try 50ppb and 1ppm first.