In icp ms nickel cones are generally used. The temperature of the plasma is between 6000- 8000k. The temperature produced by the plasma is very very higher than the melting point of Ni, then how its working ?
In icp ms nickel cones are generally used. The temperature of the plasma is between 6000- 8000k. The temperature produced by the plasma is very very higher than the melting point of Ni, then how its working ?
Hi,
you’re right, the Plasma is much hotter then the melting point of any element. So the Interface on which the cone sits is water cooled. This prevents the cones from melting.
Side fact: Most cones are made from copper with a nickel (or platinum) insert.
best Regards
Chris
Hi
That is why chiller used in such instrument..Through chiller water circulated in coil around plasma, even some instrument over detector also if peltier cooling system not used over detector..
Once chiller not working, means water circulation stopped then it will melt..
Even look like same concept used in car also to keep engine cool( u can notice during car service)
Hi Chris
Thanks, is the circulating water can bring down such a high temperature? Is the water evaporation is controlled by the fast circulation?