autosampler SPS4

Hi there, 

we have an ICP-MS 7900 and we use the autosampler SPS4. Recently we have noticed that samples have not been analysed randomly in different runs. We haven't been able see yet what happens when it skips samples. Any suggestions on what could be?  

Parents
  • Hi,

    before you do an alignment, you can check several other things. 


    1. does the sampler hit the tubes at the 4 corners in the middle? If so the alignment should be ok. 
    2. is the rack configuration correct?

    3. is the sample list altered, if so some samples could be marked as „skipped“

    4. is the sample solution completely dissolved, or could it be possible that particles block the tubing ??

Reply
  • Hi,

    before you do an alignment, you can check several other things. 


    1. does the sampler hit the tubes at the 4 corners in the middle? If so the alignment should be ok. 
    2. is the rack configuration correct?

    3. is the sample list altered, if so some samples could be marked as „skipped“

    4. is the sample solution completely dissolved, or could it be possible that particles block the tubing ??