What is equation/method that the software uses to determine the detection limits?
Find maintenance, best practice, and troubleshooting articles.
What is equation/method that the software uses to determine the detection limits?
You can check the help to read about the detection limit. Basically:
DL = 3 * standard deviation of BEC (BEC will be calculated from the calibration standards with 0 concentration).
Level 1 Conc: 0.0000 mg/L
Level 2 Conc: 0.0000 mg/L
Level 3 Conc: 0.0000 mg/L
Level 4 Conc: 0.0100 mg/L
Level 5 Conc: 0.1000 mg/L
In this example, BEC will be calculated from the first three levels. They all have zero concentration.
You can check the help to read about the detection limit. Basically:
DL = 3 * standard deviation of BEC (BEC will be calculated from the calibration standards with 0 concentration).
Level 1 Conc: 0.0000 mg/L
Level 2 Conc: 0.0000 mg/L
Level 3 Conc: 0.0000 mg/L
Level 4 Conc: 0.0100 mg/L
Level 5 Conc: 0.1000 mg/L
In this example, BEC will be calculated from the first three levels. They all have zero concentration.