Hello,
My performance report is gret. I set AutoTune before my batch. But it fails everytime. It says low sensitivity (No gas mode)
Find maintenance, best practice, and troubleshooting articles.
Hello,
My performance report is gret. I set AutoTune before my batch. But it fails everytime. It says low sensitivity (No gas mode)
Hello,
Please ensure the correct tune vial is selected and enough sample uptake/stabilization time is given in your Batch. This is in in Acq Method->PeriPump
I appreciate for your help.
How do you set preemptive Rinse?
The check box to enable this is on the bottom right of the same page. Pre-emptive rinse allows the user to select a pre meditated time before the end of the current acquisition to move the probe from sample to rinse solution, to speed up analysis time
The check box to enable this is on the bottom right of the same page. Pre-emptive rinse allows the user to select a pre meditated time before the end of the current acquisition to move the probe from sample to rinse solution, to speed up analysis time
Thank you so much!