EM pulse limit?

Hi all, 

I am trying to run the dead time calibration routine in MH for our 8900 ICP-MS, using 166/164Er tuning solutions, and after having successfully ran "EM auto setting". Everytime I try I get the following error message "EM pulse limit is incorrect. Check the system parameters." 

Has anybody seen this before? Any idea on what's going wrong?

Thanks

Romain

Top Replies

Parents Reply
  • Hi there Romaine. Thanks very much this is very helpful. Fortunately we have a (v.expensive) service contract with Agilent and we managed to get an engineer to attend to get into the engineer menu (engineer secret password only!) and reset the deadtime. However, once this was done we were still having issues and it was the same issue you had - the EM pulse limit was wildly wrong. Once set to a sensible setting all was fine. In a nutshell your helpful answer would have been a godsend if we didn't have access to a Agilent engineer and helps me understand the issue moving forward.

Children