I now have a oneNeb series 2 and it is all loose in our inert single pass chamber for the MP-AES. I don't use our inert chamber often so it is the first time this come up.
Do I need to buy some kind of fitting?
Find maintenance, best practice, and troubleshooting articles.
I now have a oneNeb series 2 and it is all loose in our inert single pass chamber for the MP-AES. I don't use our inert chamber often so it is the first time this come up.
Do I need to buy some kind of fitting?
Hi Vicky,
What type of inert spray chamber is used, maybe you need the correct o-ring for the Agilent inert spray chamber?
And is the OneNeb 2 nebulizer used before in an cyclonic glass spray chamber.
Sometimes the OneNeb 2 is fitted to tight in a glass cyclonic spray chamber when the Hexil type of connection is used. Then the OneNeb is bended a little bit inwards. Due to that the OneNeb is to small and could be to lose into the hole of the spray chamber. Maybe this the issue.
When a brand new OneNeb 2 is used and the problem isn't there then this is the issue.
If the problem is still with a brand new Agilent Inert Spray Chamber (with the correct o-rings) and brand new OneNeb 2 please contact you local sales person for advice.
cheers
Edgar
Hi Vicky,
What type of inert spray chamber is used, maybe you need the correct o-ring for the Agilent inert spray chamber?
And is the OneNeb 2 nebulizer used before in an cyclonic glass spray chamber.
Sometimes the OneNeb 2 is fitted to tight in a glass cyclonic spray chamber when the Hexil type of connection is used. Then the OneNeb is bended a little bit inwards. Due to that the OneNeb is to small and could be to lose into the hole of the spray chamber. Maybe this the issue.
When a brand new OneNeb 2 is used and the problem isn't there then this is the issue.
If the problem is still with a brand new Agilent Inert Spray Chamber (with the correct o-rings) and brand new OneNeb 2 please contact you local sales person for advice.
cheers
Edgar