Good afternoon, I would like to know if in any guide or manual indicates the calibration of the sample aspiration flow for the flame method
Find maintenance, best practice, and troubleshooting articles.
Good afternoon, I would like to know if in any guide or manual indicates the calibration of the sample aspiration flow for the flame method
The aspiration rate for a flame instrument should be 6ml/min +/- 0.5ml/min. This is measured straight out of the nebulizer, without any accessory like the SIPS or autosampler.