Good afternoon, I would like to know if in any guide or manual indicates the calibration of the sample aspiration flow for the flame method
Good afternoon, I would like to know if in any guide or manual indicates the calibration of the sample aspiration flow for the flame method
The aspiration rate for a flame instrument should be 6ml/min +/- 0.5ml/min. This is measured straight out of the nebulizer, without any accessory like the SIPS or autosampler.