I've remake all the solutions, cleaned neb and torch, new tubing. It ran last week but this week i can't get it to run
Find maintenance, best practice, and troubleshooting articles.
I've remake all the solutions, cleaned neb and torch, new tubing. It ran last week but this week i can't get it to run
I know you said you cleaned the nebulizer but do you have a brand new nebulizer available for troubleshooting?
When you cannot perform a torch scan or wavelength calibration, it either means there is something going wrong in the sample introduction area, or the pre-optics/optics has a problem. Have you checked the injector tube of the torch to ensure there is no residue or blockage?
I know you said you cleaned the nebulizer but do you have a brand new nebulizer available for troubleshooting?
When you cannot perform a torch scan or wavelength calibration, it either means there is something going wrong in the sample introduction area, or the pre-optics/optics has a problem. Have you checked the injector tube of the torch to ensure there is no residue or blockage?