ICP-MS 7700x Error:1212

  Hi.

   After the icp-ms device has been working in the plasma on position for 15-20 minutes, it gives Error: 1212 and Error: 1233 error. Forward power and refleckted power values constantly vary.

   After turning the icp-ms device into the shut down position, everything turns back to normal after removing all electrical connections and starting the device again.

   When Icp-ms moves the device to plasma on position it gives the same error again.

 

   I would be glad if you help me. Thank you.

Parents
  • Hi,

     

    it could be several parts in the icp-ms that give those errors. if you check the help and type the error number you could find the relating part of the instrument that give this error. 

    in this case the 1212 error 1233 error is relating to the 48V power supply. But be aware, most of the time more errors appear on the same time in the error list. it's important to know what error came first.

    If the sample intro parts are correct and were possible replaced already. And there is no problem with the spray chamber drain or a air leak around the spray chamber end cap (sometimes this happens with older 7700 systems) or couplings between the spray chamber and torch then there is an internal hardware problem. please contact Agilent if an internal error is the issue and give the full error list to them to figure out what's the real problem.

Reply
  • Hi,

     

    it could be several parts in the icp-ms that give those errors. if you check the help and type the error number you could find the relating part of the instrument that give this error. 

    in this case the 1212 error 1233 error is relating to the 48V power supply. But be aware, most of the time more errors appear on the same time in the error list. it's important to know what error came first.

    If the sample intro parts are correct and were possible replaced already. And there is no problem with the spray chamber drain or a air leak around the spray chamber end cap (sometimes this happens with older 7700 systems) or couplings between the spray chamber and torch then there is an internal hardware problem. please contact Agilent if an internal error is the issue and give the full error list to them to figure out what's the real problem.