HMI seems not working in agilent icp ms 7800,anyone can answer?
Find maintenance, best practice, and troubleshooting articles.
HMI seems not working in agilent icp ms 7800,anyone can answer?
It didn't work for me until I used a higher ISTD concentration (500 ppb). If you still have problems with drift, aspirate a heavy-matrix sample for a few minutes to coat the cones. I keep a "dirty" set of cones available for when I want to use HMI.