7900 pause in batch sampler position

After a sample is completed, there is an "automatic" rinse done at cups 1 and 2. If we have a pause in the batch then the sample probe sits in cup 2.  We would like for it to go to the home position then pause.

 

Is there a script to do this, or a software change in the "automatic" rinse routine to end in the home position before going to the next line?

Parents
  • Hi,

     

    There is a way to park but we need more details.

    What is the hardware and software setup/version like:

    - autosamper

    - isis

    - software version

    - type of matrix

     

    In ICP-MS Masshunter 4.5 you can set the last "rinse step" line into the home position for example one second. Then always the run ends in the home position. Picture below is different when ISIS is used.

     

     

Reply
  • Hi,

     

    There is a way to park but we need more details.

    What is the hardware and software setup/version like:

    - autosamper

    - isis

    - software version

    - type of matrix

     

    In ICP-MS Masshunter 4.5 you can set the last "rinse step" line into the home position for example one second. Then always the run ends in the home position. Picture below is different when ISIS is used.

     

     