Re: Can I use Gold Chloride when avoiding mercury memory effects when doing ICP-MS analysis?
tmcsweeneyJan 30, 2019 5:47 AM (in response to mivuyo12)
2 people found this helpfulYes Gold helps with keeping Hg in solution and its wash out.
HCl also helps with wash out and keeping Hg in solution
-
Re: Can I use Gold Chloride when avoiding mercury memory effects when doing ICP-MS analysis?
mivuyo12 Jan 31, 2019 12:12 AM (in response to mivuyo12)What is the best concentration to use? Do I have to add it on wash vails not the original samples and calibration samples?
-
Re: Can I use Gold Chloride when avoiding mercury memory effects when doing ICP-MS analysis?
tmcsweeneyJan 31, 2019 5:21 AM (in response to mivuyo12)
1 person found this helpfulwe
use 0.5% hcli
use in all my rinsesas
for gold concentration i’m not exactly sure as that can vary-
Re: Can I use Gold Chloride when avoiding mercury memory effects when doing ICP-MS analysis?
mivuyo12 Feb 14, 2019 6:34 AM (in response to tmcsweeney)So do I use 0.5% HCL as a rinse solution? So is it better to use this solution after every 5 samples ran?
-
Re: Can I use Gold Chloride when avoiding mercury memory effects when doing ICP-MS analysis?
tmcsweeneyFeb 14, 2019 1:31 PM (in response to mivuyo12)
Yes you can use 0.5% HCl as the rinse solution.
-
-
-