    Making my Laboratory Control Spectrum the "blank" for my samples

    tylerfunke

      Hey Agilent Community,

      I am very new to Masshunter software and had a few questions regarding my approach to my untargeted metabolomics.

      I have been trying to make my laboratory control chromatogram my "blank" basically, and then trying to see which peaks in my samples are different from the control. Is there a method available to easily do this? I feel like I have been having more trouble than I should - as it doesn't seem like that complex of a question. I have looked into possibly creating a custom PCDL that has the control compounds found through the "find by molecular feature" or "find by auto MS/MS" tool, and I have also thought that maybe it would be possible to just background subtract the entire control spectrum from my samples. My data is MS and MS/MS collected from a QTOF, if that is relevant.

      Does anyone have suggestions on how I should go about answering this question? Or maybe could someone point me in the direction of a webinar or guide that could help me with this question? Again, I am pretty new to using Masshunter so I apologize if this question seems basic or has already been answered.

        • Re: Making my Laboratory Control Spectrum the "blank" for my samples
          hexatriene

          Hey tylerfunke,

           

          It sounds like you're trying to re-create our MassHunter Mass Profiler software! This is exactly what it is designed to do:

          MassHunter Software: Mass Profiler | Agilent

           

          Here's the quick start guide: https://www.agilent.com/cs/library/usermanuals/public/G3297-90017_MassProfiler_QuickStart.pdf

           

          ~Hexatriene

            • Re: Making my Laboratory Control Spectrum the "blank" for my samples
              tylerfunke

              While the Mass Profiler looked to be what I was trying to do, I do not believe I have that software... I have the 2011 B.05.00 version and could not find a disc for the mass profiler. Is there a possibility I could do something similar with just the Quantitative Analysis software and the PCDL?

                • Re: Making my Laboratory Control Spectrum the "blank" for my samples
                  hexatriene

                  Hi tylerfunke, unfortunately Qual does not have a concept of a "Blank" data file. It cannot perform the required 'alignment' necessary to ensure that two features which share the same mass are the same compound - and therefore has no mechanism to remove a set of compounds from another set. If restricted entirely to Qual - I would do this by exporting the feature lists found in each data file - right-click on the compound list - and export to Excel. Then do my workup in excel to:

                   

                  1.) Align the compound lists

                  2.) Find differences between them

                   

                  There is separate tool called Profinder (which is free) which is capable of recursive 'find by molecular feature'. This will use a feature found in one data file to more thoroughly search for that feature in other data files (and supports chromatographic warping (to fix retention time shifting issues)). Profinder ships on the LCMS Supplemental Disk (version B.09, part number G3336-60108) which you can download from agilent.subscribenet.com. If you're having trouble accessing agilent.subscribenet.com - you can call our online technical support number (Contact Us | Agilent, option 3, then 2) and request free media replacement (or a download link for the same).

                   

                  Profinder does not do fold change analysis. You would still need to do that part by hand - but the alignment should be taken care of for you. Profinder, like Qual, produces a compound list which can be exported to excel for you to do your own further analysis. This is basically the same workflow I described above for Qual - but Profinder is geared for cross-datafile feature finding. The real Agilent solution for this problem is to use Profinder to do the feature finding and alignment, then use Mass Profiler or Mass Profiler Professional to do the statistical and fold-change analysis.