Re: Filtering of Blank Sheets of the Report
ryoboyleNov 26, 2018 7:52 AM (in response to mr.white)
Hello,
Would you mind providing more information about your system and what you are attempting? Is this OpenLab CDS 2.x software or another product? Is the report template in question a single injection report or a sequence summary report?
-
Re: Filtering of Blank Sheets of the Report
mr.white Nov 27, 2018 1:00 AM (in response to mr.white)Open Labs 2.1. Its a single Injection Report.
I filtered the SST Out, But the other pages remain blank.
Thank you
-
Re: Filtering of Blank Sheets of the Report
ryoboyleNov 27, 2018 7:13 AM (in response to mr.white)
1 person found this helpfulHello mr.white,
For single injection reports, if a data file is unable to populate the fields on the report (due to filters, no peaks, etc...) then the report will show up as a blank page. If you have a report that is specific for calibration samples, then the only way to avoid blank pages is to only use it for your calibration sample data files. A possible solution would be to link a different method to your calibration data files than your sample data files. Each processing method would have an appropriate report specified for that sample type.
Another possible solution would be to use a sequence summary report.
-
Re: Filtering of Blank Sheets of the Report
hongqiNov 28, 2018 5:42 PM (in response to mr.white)
1 person found this helpfulHi,
Maybe you can use sequence summary report. you can group by injection ID, then apply filter on calibration sample in the group.
-
-
Re: Filtering of Blank Sheets of the Report
mr.white Dec 4, 2018 12:06 AM (in response to mr.white)Thank you for the help :-)
I will mark it as correct as soon as I have time to validate it.
I just need a bit more time