This Information Applies to: Agilent AriaMx real-time PCR system
Issue
Motor calibration (MC) is a procedure that identifies the location of each well scanned by the optical modules for each slot.
Background
This calibration is completed before the instrument is shipped, so you usually do not need to do it again. However, if the instrument’s touchscreen shows a notification that motor calibration is required (as shown in Figure 1), click OK and follow the on-screen instructions.
Figure 1. AriaMx raw data plots.
1. Notification when motor position is not calibrated
If Agilent technical support asks you to perform a motor calibration, please perform the following procedure.
Tip: Adding and removing optical modules does not cause the instrument to require a motor calibration.
|
Items required
Steps to Follow
Access Settings:
(
Figure 2).
Initiate Motor Calibration:
Prepare the plate:
Configure Optical Modules:
Follow on-screen instructions:
Perform background Calibration:
Tip: After performing the motor calibration, it is crucial to verify the motor calibration values using the AriaMx support tool. Before starting the motor calibration, download our AriaMx Support Tool software here. You can find here the Aria Support Tool User Guide. Once you have run the motor calibration, verify the motor calibration values by following the protocol at page 9 of the Aria Support Tool User Guide.
|
Tip: View Performing a Background Calibration on the AriaMx |
earn how to effectively operate your Agilent AriaMx:
AriaMx Real-Time PCR System Setup and User Guide on Agilent.com Aria Real-Time PCR Software User Manual on Agilent.com |