This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx real-time PCR electronic tracking (ET) software
Issue
Backing up audit logs from AriaMx ET software
Background
When you need to move your AriaMx ET software installation from one computer to another, it is crucial to transfer the AriaMx audit logs as well. Unfortunately, there is no direct way to transfer the audit logs within the AriaMx software itself. While administrators can export the audit logs to Excel, importing them back is not supported.
Find more details about this limitation in chapter 16 of the AriaMx Real-Time PCR Software User Manual.
|
To back up and restore your data, you can use SQL Server Management Studio.
Steps to Follow
Caution: Customers are responsible for verifying that the backup and reloaded database is complete.
|
For more detailed instructions, see the PDF How to Backup 2019 SQL Server (For AriaMx Electronic Tracking Users). |
Learn how to effectively operate your Agilent AriaMx real-time PCR system:
AriaMx Real-Time PCR System Setup and User Guide on Agilent.com AriaMx Real-Time PCR software User Manual on Agilent.com |
For Research Use Only. Not for use in diagnostic procedures.