This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx real-time PCR software
Issue
AriaMx software automatically changes the values of standard solutions.
Background
When setting quantity values for standard wells on the Plate Setup screen, select the Units (for Plate) (e.g., micrograms, nanograms) first, then enter the Amount. If you change the units after entering a value, the AriaMx software will automatically convert the amount to maintain the same concentration.
For example, if you enter 500 nanograms and then select picograms as the unit, the software will convert it to 500,000 picograms (Figure 1). To ensure that the values remain consistent, select the units for the replicate set of standard wells first, then enter the starting amount.
Figure 1. Automatic conversion of the value in the Amounts field in the AriaMx software.
The change in the Units (for Plate) drop-down menu is indicated with a red arrow. The resultant change in the Amount field is highlighted with a red box.
Steps to Follow
The workflow for assigning replicate numbers and starting amounts to standard wells is described in the following steps and shown in Figure 2.
Figure 2. Workflow of assigning replicate numbers and a starting amount to standard wells in the plate setup.
Caution: Without correct values, the absolute quantification may be incorrect.
|