Chromatography Software Articles How to Export OpenLab System Activity Log and Diagnostics Logs for Troubleshooting
  • Chromatography Software Articles
  • Chromatography Software Videos
  • Chromatography Software Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Chromatography Data Systems
    • +OpenLab CDS Administration
    • +OpenLab CDS Acquisition
    • +OpenLab CDS Data Analysis
    • -OpenLab CDS Troubleshooting
      • OpenLab CDS Online Signals Partial or Missing Display During Acquisition
      • How to restart Agilent OpenLab Services in Windows 10
      • How to Export OpenLab System Activity Log and Diagnostics Logs for Troubleshooting
      • General Procedure and Common Issues with Licensing of Agilent OpenLab CDS
      • OpenLab ChemStation Not Starting - Basic Troubleshooting

How to Export OpenLab System Activity Log and Diagnostics Logs for Troubleshooting

Created by Taichiro Hayashi Taichiro Hayashi over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 3 months ago

This Information Applies To: OpenLab CDS ChemStation (version C.01.xx), OpenLab CDS EZChrom (version A.04.xx), OpenLab CDS (version 2.x), OpenLab ECM XT


Issue:

An Agilent Support Engineer may request system internal log files when troubleshooting Agilent OpenLab Workstation, Client, or Instrument Controller. The log files include the OpenLab System Activity Log and OpenLab Diagnostics Log.


System Activity Log Steps to Follow:

  1. Access the OpenLab Control Panel by clicking either the Control Panel or Agilent OpenLab AB Control Panel icon. The icon varies with the version. 

  2. Select Administration > System Activity Log. The System Activity Log pane with appear. 

  3. In the System Activity Log enter fields for User, Level, Subsystem, Description to filter the data. The filter values will be supplied by the Agilent Support Engineer. The From/To field must include the entire period in which the problems have been experienced. 

  4. This image is an example of filter values:

  5. Select Print > Print Detailed to open the System Activity Log.

  6. Select Export > PDF.

  7. Specify the destination location, enter a File name (the default is Report.pdf) and click Save.

  8. The report can now be emailed to the Agilent Support Engineer.

System Diagnostic Log Steps to Follow:

  1. Access the OpenLab Control Panel by clicking either the Control Panel or Agilent OpenLab AB Control Panel icon. The icon varies with the version. 

  2. Select Administration > Diagnostics. The Management toolbar will appear.

     

  3. There are two options for log files on the toolbar:
    1. Server Log Files: Use this when your instrument is part of a client/server system. Use options in the Server Log Files pane.
    2. Local Log Files: Use this when your instrument is a standalone system. Use options in the Local Log Files pane.

      In the appropriate pane, click Select All > Save Logs

  4. Specify the destination location, enter a File name, and click Save.

  5. The report can now be emailed to the Agilent Support Engineer.

 Tip 
Learn how to effectively operate and troubleshoot your Agilent OpenLab System:
Agilent OpenLab CDS Tutorials at Agilent.com
  • openlab ezchrom
  • OpenLab Chromatography Data System
  • Troubleshooting
  • OpenLab CDS
  • OpenLab ChemStation
  • openlab ecm xt
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help