Mass Spectrometry Software Articles No Data Points Error in MassHunter Quantitative Data Analysis
  • Mass Spectrometry Software Articles
  • Mass Spectrometry Software Videos
  • Mass Spectrometry Software Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -MassHunter Software
    • +MassHunter Acquisition
    • +MassHunter Qualitative Data Analysis
    • +MassHunter Quantitative Data Analysis
    • -MassHunter Troubleshooting
      • How to Reset the Agilent GC/MSD Tune File to Default Values
      • No Data Points Error in MassHunter Quantitative Data Analysis
      • Out of Memory Warning on MassHunter
      • How To Configure a New GC/MS with the Agilent GC/MS Configuration Editor
      • Error Message 'The ChemStation did not Complete its Initialization Cycle' when Starting MassHunter GC/MS Acquisition Software
    • +MassHunter Tips

No Data Points Error in MassHunter Quantitative Data Analysis

Created by Xu Li-hua Xu Li-hua over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 10 months ago

This Information Applies To: Agilent MassHunter Quantitative Data Analysis Software


Issue:

The "NO DATA POINTS" error message appears in the Compound Information window in the lower-left corner in Quantitative Analysis. Sometimes, this problem may also be described as 'not seeing any chromatogram' or 'no peaks are visible in the chromatogram'. In this situation. A possible reason for this error is the MSTree2.bin file in the data file may be corrupted. To solve this error, see the following procedure.


Steps to follow:

  1. Close MS Quantitative Analysis.

  2. Using the File Explorer, go into the directory where the data files where the issue is occurring, are found. Take the data Y01904017.D as an example - its MSTree2.bin file is located in the folder Y01904017.D / AcqData (see Figure 1). There should be one MSTree2.bin file for each data file/folder in the directory.


    Figure 1. AcqData folder files detail
    1 - MSTree.2 bin file path  2 -  MSTree2.bin file 
  3. Delete the MSTree2.bin file for the corrupted data.
  4. Re-open Quantitative Analysis
  5. Reload the batch. A new MSTree.bin file will be created for each data file.
  6. Reanalyze the batch.

If the preceding procedure does not solve the issue, contact your local Agilent support representative.

 Tip 
Learn how to effectively operate your MassHunter software:
MassHunter Quantitative Analysis for GC/MSD Familiarization Guide user manual on Agilent.com.
  • Troubleshooting
  • gcms data analysis
  • Quantitative Data Analysis Software
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help