Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Mass Spectrometry Software Portal
  • Agilent Knowledge Portal
Mass Spectrometry Software Portal
Mass Spectrometry Software Articles How to Create and Run a Sequence in GC/MS MassHunter Data Acquisition Software
  • Mass Spectrometry Software Articles
  • Mass Spectrometry Software Videos
  • Mass Spectrometry Software Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -MassHunter Software
    • -MassHunter Acquisition
      • How to Run a Single Sample with ALS in Agilent GC/MS MassHunter Data Acquisition Software
      • How to Create and Run a Sequence in GC/MS MassHunter Data Acquisition Software
    • +MassHunter Qualitative Data Analysis
    • +MassHunter Quantitative Data Analysis
    • +MassHunter Troubleshooting
    • +MassHunter Tips

How to Create and Run a Sequence in GC/MS MassHunter Data Acquisition Software

Created by Xu Li-hua Xu Li-hua 11 months ago | Last modified by Carlos Vargas Carlos Vargas 6 months ago

This Information Applies To: Agilent GC/MS MassHunter Data Acquisition Software


Issue

Multiple samples can be run without user intervention using an Automated Liquid Sampler (ALS) in combination with a Sequence.

Users can perform simple edits to a new or existing sequence template to specify the sequence of vial numbers, sample names, data file names, and storage locations, etc. The sequence can then be run according to the following procedure. 


Steps to follow

The video shown in Figure 1 demonstrates the following steps on how to create a sequence:

1. Select Sequence > Edit Sequence or click the Edit Sequence  icon on the Instrument Control panel to display the sequence table. Each line in the sequence table contains information for the analysis of one sample (one vial for an ALS).

2. To use a previously saved sequence table column layout, click Tools and select Load Column Layout from the drop-down list to display the Open dialog box. Select the file that contains the correct layout and click Open.

3. If you want to add more samples than the current table can hold, click the  drop-down list, and select 1 Sample, 5 Samples, and N Samples to add one, 5, or customized numbers of new sample lines in the table.

4. Edit the sequence table by entering the required values for all parameters that apply to each sample.

Tip  
Tips: When using the 7693 ALS with a G4515A Barcode Reader tray installed, you must enter a value in the Expected Barcode column. If you are not entering a barcode value, enter the non-case sensitive string Off to disable reading of a non-existent barcode.
To use Sample Overlap on an ALS and tray, the sequence must use a single method defining Sample Overlap for all samples.

5. To copy the value in a selected cell down to all subsequent cells, select a cell in the table and click .

6. To increment up a number's value in a selected cell to all subsequent cells, select a cell in the table, and click .

7. When you are finished, click OK to close the sequence table.

8. Save the sequence.

 
Figure 1. A video demonstrating how to create a sequence in GC/MS MassHunter Data Acquisition

9. Select Sequence > Run Sequence or click the Run Sequence  icon on the Instrument Control panel. The Start Sequence dialog Basic tab is displayed in Figure 2.

 
Figure 2. The Start Sequence dialog Basic tab

10. Specify the Method Sections to run, Sequence Barcode Options, and any Pre-Sequence or Post-Sequence Macros/Command needed. Optionally, click the Recurring Keyword tab to automatically run a recurring keyword based on a specified interval.

11. Click Run Sequence to start processing samples. A sequence status bar is displayed above the Instrument Control panel for monitoring and controlling the sequence (see Figure 3).

 
Figure 3. Sequence running status

 Tip 
Learn how to effectively operate your Agilent MassHunter Software:
Agilent 5977 GC/MSD with MassHunter Workstation
Agilent 7000 or 7010 Triple Quadrupole GC/MS with MassHunter Workstation e-learning paths available from Agilent education
GCMS-MH-2120e - GC/MS MassHunter Sequencing and the Sequence Table Assistant (Ver. B.07)
GCMS-MH-2121e - GC/MS MassHunter: Importing Simulating and Running a Sequence (Ver. B.07)  e-learning courses available from Agilent education
  • Operation
  • MassHunter Acquisition Software
  • GCMS Data Acquisition
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help