Mass Spectrometry Software Articles How to Run a Single Sample with ALS in Agilent GC/MS MassHunter Data Acquisition Software
  • Mass Spectrometry Software Articles
  • Mass Spectrometry Software Videos
  • Mass Spectrometry Software Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -MassHunter Software
    • -MassHunter Acquisition
      • How to Run a Single Sample with ALS in Agilent GC/MS MassHunter Data Acquisition Software
      • How to Create and Run a Sequence in GC/MS MassHunter Data Acquisition Software
    • +MassHunter Qualitative Data Analysis
    • +MassHunter Quantitative Data Analysis
    • +MassHunter Troubleshooting
    • +MassHunter Tips

How to Run a Single Sample with ALS in Agilent GC/MS MassHunter Data Acquisition Software

Created by Xu Li-hua Xu Li-hua over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 7 months ago

This Information Applies To: Agilent GC/MS MassHunter Data Acquisition Software


Issue

Preparing and running a sample is essential for data acquisition. The steps below detail how to run a single sample using an Automatic Liquid Sampler (ALS) on your GC/MS system.


Steps to follow

The following steps to run a single sample using a previously created method are demonstrated in the video shown in Figure 1.

1. Browse and select the method you wish to use.

2. Select Method > Run Method or click the large green arrow on the Instrument Control panel to display the Start Run dialog box.

3. In the Start Run dialog box, enter a Data File Name.

4. Enter the Sample Name and Misc. Info., which is optional information to be stored with your data file.

5. Enter a Sample Amount, Multiplier, and the Vial Number location where the sample is stored in the ALS tray.

  Tip
Tips: When using the 7693 ALS with a G4515A Barcode Reader (BCR) tray installed, you must enter a value in the Expected Barcode field. If you are not entering a barcode value, enter the non-case sensitive string Off to disable reading of a nonexistent barcode.

6. In the Methods Section to Run box, ensure that Data Acquisition is selected. If you select the Data Analysis(MassHunter DA) which is optional and will start the quantitative data analysis program with the data file acquired after the run.

7. Click OK and Run Method.

Once you submit the single sample run, the vial number, sample name, and data file name are displayed in the sample run box and the run time is displayed above the total run time in the Run Time box.

The status box at the top-left corner shows the system in Pre-Run status, which changes to Run status after the sample is injected.

The Instrument Control panel icons are grayed out and all instrument controls except the Stop Run icon are disabled until the run is over.

 
Figure 1. How to run a single sample using a previously created method

 Tip 
Learn how to effectively operate your Agilent MassHunter Software:
Agilent 5977 GC/MSD with MassHunter Workstation
Agilent 7000 or 7010 Triple Quadrupole GC/MS with MassHunter Workstation e-learning paths available from Agilent education
  • Operation
  • MassHunter Acquisition Software
  • GCMS Data Acquisition
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help