Mass Spectrometry Software Articles Using the Calibration Curve Window in MassHunter Quantitative Analysis Software
  • Mass Spectrometry Software Articles
  • Mass Spectrometry Software Videos
  • Mass Spectrometry Software Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -MassHunter Software
    • +MassHunter Acquisition
    • +MassHunter Qualitative Data Analysis
    • -MassHunter Quantitative Data Analysis
      • Creating a New Batch in MassHunter Quantitative Analysis
      • Creating a Method in MassHunter Quantitative Analysis – Quant-My-Way user interface
      • Calculating the Final Concentration in MassHunter Quantitative Analysis
      • How to Average Calibration Responses in MassHunter Quantitative Analysis
      • Outlier Setup in MassHunter Quantitative Analysis Software
      • Combine multiple Methods in MassHunter Quantitative Analysis into a Single Method
      • Using the Compound Information window in MassHunter Quantitative Analysis Software
      • Modify the Integration Parameters in MassHunter Quantitative Analysis
      • Using the Batch Table in MassHunter Quantitative Analysis Software
      • Using the Calibration Curve Window in MassHunter Quantitative Analysis Software
    • +MassHunter Troubleshooting
    • +MassHunter Tips

Using the Calibration Curve Window in MassHunter Quantitative Analysis Software

Created by Xu Li-hua Xu Li-hua 11 months ago | Last modified by Carlos Vargas Carlos Vargas 6 months ago

This Information Applies To: Agilent MassHunter Quantitative Analysis Software


Issue

The Calibration Curve window tools help with setting an appropriate calibration curve for each compound in the calibration table.


The Calibration Curve window displays the following information for the current compound:

  • Calibration points with curve fit
  • Curve fit type, origin, and weighting
  • Curve fit equation and metrics
  • Number of levels and levels used
  • Number of points and points used
  • Optionally QC, CC, and ISTD points.

Steps to follow

  1. View the Calibration Curve window. If the Calibration Curve window does not appear in the layout, click View > Calibration Curve.

  2. Change the Scale and Zoom plot: 

    Autoscale the full data range click Turn On/Off auto scale in Calibration pane menu bar button.
    Zoom either or both the horizontal and vertical scale right click and drag the area to be zoomed on the calibration plot.
    Adjust just the horizontal or vertical scale range right click and drag the scale up or down.
    Adjust just the horizontal or vertical scale position left click and drag the scale up or down.
    Undo the horizontal or vertical zoom or scale position click the Auto Scale X-Axis or Auto Scale Y-Axis menu bar button. 

    See this video:

  3. Navigate between compounds

     In the Calibration Curve menu bar, click left or right blue arrow menu bar buttons 

  4. Change calibration curve type.

     In the Calibration Curve menu bar, click Type and select from the drop-down list.

  5. Change calibration Origin type

    In the Calibration Curve menu bar, click Origin and select from the drop-down list.

  6. Change the weighting of calibration data points in the calibration curve

    In the Calibration Curve menu bar, click Weight and select from the drop-down list

     Tip 
    Tip: When a large calibration range is used, the lower concentration calibration accuracy (residuals) may be improved by selecting a 1/x or 1/x^2 weight rather than the default of None.
  7. Display the ISTD response, QC, and CC points

    In the Calibration Curve menu bar, click ISTD of the current compound for each data point of the calibration curve (applicable only to ISTD calibration methods).
    In the Calibration Curve menu bar, click QC.
    In the Calibration Curve menu bar, click CC.

  8. Curve Fit Assistant

    Right click the calibration plot and enable the Curve Fit Assistant. 
    Click a Curve Fit table row to display the alternate curve fit on the calibration plot (zoom select areas of the plot to further assess). Review alternate settings for Type, Origin, Weight and # of Disabled Points to select the settings specified in your calibration protocol. Alternatively, evaluate alternate settings for improved calibration curve fit.
    Right click the calibration plot and select Accept Assistant Curve if you decide to use the currently viewed alternate curve fit.
    Right click the calibration plot and disable the Curve Fit Assistant to close the Curve Fit table. 

     Tip  Tip: The Curve Fit Assistant table can be:
    - Sorted by any of the column headings. Left click the column heading. 
    - Filtered by one or more column values. Left click the column heading filter drop-down menu.

    The following video demonstrates the operation of Curve Fit Assistant.

Tip

  

Learn how to effectively operate MassHunter software:
Agilent 5977 GC/MSD with MassHunter Workstation e-learning path available from Agilent education
GCMS-MH-2150scV3 - Agilent MassHunter Quantification for Single Quadrupole GC/MS with Cloud Laboratory (Version 10.1) e-learning course available from Agilent education

  • Operation
  • GC/MS Data Analysis
  • Quantitative Data Analysis Software
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help