Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Liquid Chromatography Portal
  • Agilent Knowledge Portal
Liquid Chromatography Portal
LC Articles Cleaning Recommendations for Agilent ELSDs
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • -LC Maintenance
      • +LC Pumps Maintenance
      • +LC Injectors Maintenance
      • -LC Detectors Maintenance
        • Removing and Re-Installing Visible and UV lamps in Agilent 1260 Infinity G1315x DAD or G1365x MWD
        • Optical Balance Procedure for the Refractive Index Detector
        • Cleaning Recommendations for Agilent ELSDs
    • +LC Troubleshooting
    • +LC Best Practices
    • +LC Firmware

Cleaning Recommendations for Agilent ELSDs

Created by Xu Li-hua Xu Li-hua over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 6 months ago

This Information Applies To: Agilent ELSDs G4260A, G4260B, G4261A, G4621B, G7102A, G7102B


Content:

If the evaporator tube of ELSD becomes contaminated with non-volatile material resulting in poor chromatography, please use the steps below to troubleshoot and clean the ELSD.


Steps to Follow:

It is recommended that the instrument is initially washed with a solvent suitable for contamination. 

  Caution
Caution: Do not use solvents that contain additives when performing the cleaning procedure.

 Caution 
Caution: Please read all relevant MSDS for the solvents being used, and follow all safety guidance including storage, handling, ventilation and extraction requirements, and the use of appropriate PPE.

One of the “cleaning” solvents recommended is Acetone/Water (50/50) (based on your application, select the appropriate solvent to clean the contamination).

Cleaning Procedure: 

  1. Set Evaporator temperature to 40 ℃.
  2. Set the Nebulizer temperature to 40 ℃.
  3. Set gas flow to 2.5 SLM.
  4. Connect the ELSD directly to the HPLC pump (bypassing the ALS and column).
  5. Pump the “cleaning” solvent into the instrument at 1 or 2 mL/min (while in the RUN mode).
  6. Perform this procedure overnight or for a minimum of 4 hours. Make sure that there is the required volume of cleaning solvent for the flushing time. 
  7. During the cleaning procedure, the ELSD response may show a full-scale deflection. This indicates that contamination is being flushed from the detector.
  8. Following the cleaning procedure, reconnect the column and ALS and inject a sample of known response.
  9. Repeat the wash procedure with a different solvent, or for a longer period if the ELSD signal is not improved.

Preventative Maintenance:

Depending on use, it is recommended to clean the evaporator tube at least once a week or every 40 hours.

The build-up of contamination can be identified early by monitoring the baseline noise over time, or by checking the ELSD offset reading.

To check the ELSD offset value;

  1. Switch off the ELSD.
  2. Switch the detector back on and place it into RUN mode.
  3. Record the ELSD reading from the front panel -this is the OFFSET value
  4. Offset values above 140 mV indicate contamination and it is recommended the ELSD be cleaned.

It is also recommended to clean the unit following the use of buffers, or where the sample matrix contains impurities (i.e. "dirty" sample).

If cleaning the unit does not cure the problems then contact Agilent Technologies for further assistance.

 

 Tip
Learn how to effectively maintain your Agilent ELSD:
ELSD User Manual on Agilent.com
  • Detectors
  • elsd
  • 1290 infinity
  • Maintenance
  • 1260 Infinity
  • 1260 Infinity II
  • 1200 Series
  • Analytical LC
  • 1290 Infinity II
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help