LC Articles How to Access Diagnostic Buffer File from LC Modules for Advanced Troubleshooting
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • +LC Maintenance
    • -LC Troubleshooting
      • Troubleshooting High Pressure Issues in Agilent LC Systems
      • Troubleshooting Peak Area Variation Between Injections
      • Forced Cold Start Procedure
      • Troubleshooting Remote/Cluster Partner Missing Error
      • How to Access Diagnostic Buffer File from LC Modules for Advanced Troubleshooting
      • How to Free up Licenses After License Check Message on Agilent Lab Advisor Software
      • "Could not open channel RD" error message in Lab Advisor
      • +LC Pumps Troubleshooting
      • +LC Injectors Troubleshooting
      • +LC Column Compartments Troubleshooting
      • +LC Detectors Troubleshooting
    • +LC Best Practices
    • +LC Firmware

How to Access Diagnostic Buffer File from LC Modules for Advanced Troubleshooting

Created by Max Schütte Max Schütte over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 7 months ago

This Information Applies To: All Agilent LC modules.


Issue:

If there are troubleshooting issues with LC hardware as a part of an Agilent OpenLab system, Agilent Support representatives may ask you to provide them with a Diagnostic Buffer file to help solve the issue. This is a .txt file that includes additional information for all LC modules, PC hardware, and software in addition to the information shown in the System Report. Please follow this procedure to create, save, and send the file to your local Agilent Support representative.


Steps to follow:

 Tip  If Agilent Lab Advisor software is not installed, or if your current system is not added in Lab Advisor, then please see the following articles for details on how to Install Lab Advisor and How to add your system.

Open Agilent Lab Advisor software, connect to your system using the Connect button, and follow these steps (Figure 1):

  1. Under Service & Diagnostics, select the module to export diagnostic buffer information.
  2. Select the option Diagnostic Buffers and press the Run tab (select No in the Trace Wizard Window).
  3. Click the Read Buffers tab (at the top) and wait until it is finished.
  4. Click the Write to File tab (at the top) and name your text file with your company name and instrument name. Add instrument actuals should be included.
  5. Click Save. A message appears that the file has been successfully created. Locate it and send it via email to your local Agilent Support representative.


Figure 1.
 How to create the Diagnostic Buffer file

 Tip 
Tips: Find the Lab Advisor User Manual here.

 Tip 
Learn how to effectively troubleshoot your Agilent HPLC with OpenLab CDS:
Agilent 1260 Infinity LC Systems with OpenLab CDS e-learning course available from Agilent education
  • Troubleshooting
  • Lab Advisor
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help