Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Liquid Chromatography Portal
  • Agilent Knowledge Portal
Liquid Chromatography Portal
LC Articles How to Sonicate the Inlet and Outlet Valves of Agilent 1100, 1200 and 1260 Series Pump Heads
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • -LC Maintenance
      • -LC Pumps Maintenance
        • How to Replace the PTFE Frit of the Purge Valve in LC Pumps
        • How to Replace an Inlet Valve in a G7120A Agilent 1290 Infinity II High Speed Pump
        • How to Change the Passive Inlet Valve (PIV) on a 1260 Isocratic or Quaternary Pump
        • How to Change the Outlet Valve on a 1260 Isocratic or Quaternary Pump
        • How to Remove the Pump Head From an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump
        • How to change seals in a 1260 Infinity Binary, Quaternary, or Isocratic Pump without Seal Wash Option
        • Reinstalling a pump head on Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules
        • Seal Wear in Procedure for 1260 Isocratic and Quaternary Pumps
        • How to Sonicate the Inlet and Outlet Valves of Agilent 1100, 1200 and 1260 Series Pump Heads
      • +LC Injectors Maintenance
      • +LC Detectors Maintenance
    • +LC Troubleshooting
    • +LC Best Practices
    • +LC Firmware

How to Sonicate the Inlet and Outlet Valves of Agilent 1100, 1200 and 1260 Series Pump Heads

Created by Paola Labarbuta Paola Labarbuta over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 7 months ago

This Information Applies To:  Agilent 1100, 1200, 1260 LC Pumps.


Issue:

The following procedure allows the cleaning of residue from the pump's check valves, often caused by polymerization of aged acetonitrile or salt-based solutions that may prevent the inlet and outlet valves from working correctly.

When inlet and/or outlet valves do not close correctly, the pump pressure will fluctuate. When monitoring the pressure trace, the user may see a sawtooth pattern (up and down) of various amplitudes.

If the sonication of the valves does not solve the issue, please contact Agilent Technologies for further support.


Steps to follow:

  • Switch off the pump at the main power switch.

  • Open the doors of the module.

  • Disconnect the outlet valve.

    The outlet valve is positioned above the pump head as depicted in Figure 1.

    1. Disconnect the valve capillary.

    2. Unscrew the valve by turning counterclockwise, using a wrench.


Figure 1. Outlet Valve

  • Disconnect the inlet valve

The inlet valve is positioned below the pump head as shown in Figure 2. It shows an Active Inlet Valve.


Figure 2. Active Inlet Valve

The inlet valve can be either passive or active.

To disconnect the Active Inlet Valve, follow these steps:

  1. Disconnect the valve cable by pulling it straight out of the connector. Do not twist and unscrew the valve cable from the connector as you will expose the electrical pins that can be damaged.

  2. Disconnect the Solvent Inlet Tube.

  3. Unscrew the valve using a 14 mm wrench.

  4. Carefully remove the cartridge from the valve body (see Figure 3) being careful not to damage the edges as that help to form the seal.


aiv CARTRIDGE

Figure 3. Active Inlet Valve Cartridge.

The Passive Inlet Valve can be disconnected similarly to the outlet one, and it will not need to be further disassembled before sonication.

  • Sonicate Valves for 30 minutes

    Place the Active Inlet Valve Cartridge/ Passive Inlet Valve and the Outlet Valve in Methanol or Isopropanol and sonicate them for 30 minutes. If it's suspected that salts may have precipitated in the valves, then use warm water instead of an organic solvent to sonicate. Do not sonicate the Active Inlet Valve body as it will destroy the electronic components.

    Then wash them out with water and re-assemble them on the pump head. See your pump modules user manual for detailed instructions on re-assembly.

  • After the valve sonication procedure, it is suggested to purge the pump with water at 60 degrees, keeping the purge valve open.

    Wash the system with isopropanol (excluding the column).

    Finally equilibrate the system with your solvents before use.

 Tip 
Learn how to effectively troubleshoot your HPLC pump:
HPLC-0GEN-2201e - Chromatographic Troubleshooting for HPLC e-learning course available from Agilent education
  • 1100 series
  • Pumps
  • Maintenance
  • 1260 Infinity
  • 1260 Infinity II
  • 1200
  • Analytical LC
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help