Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Liquid Chromatography Portal
  • Agilent Knowledge Portal
Liquid Chromatography Portal
LC Articles How to Autoreference the G7167A/B Infinity II Multisampler
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • -LC Maintenance
      • +LC Pumps Maintenance
      • -LC Injectors Maintenance
        • Exchanging the Needle Assembly on Agilent G1367A-D Autosamplers
        • How to Change the Needle on a G7167x Multisampler
        • How to Change the Needle and Needle Seat in an Agilent G7129x Vialsampler
        • How to change the needle seat on a G7167x Multisampler
        • How to Autoreference the G7167A/B Infinity II Multisampler
        • How to replace the peristatic pump on 1290 Infinity (G4226x) and High Performance Autosamplers (G1367x)
      • +LC Detectors Maintenance
    • +LC Troubleshooting
    • +LC Best Practices
    • +LC Firmware

How to Autoreference the G7167A/B Infinity II Multisampler

Created by Sarah Scheirs Sarah Scheirs over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 6 months ago

This Information Applies To: Agilent G7167A/B Infinity II Multisamplers


Issue:

The Agilent Multisampler autoreferencing procedure uses predefined positions on the base plate and the sample hotel to calibrate the positioning of the needle parkstation and the sample hotel. Auto referencing is required to compensate for deviations in positioning the needle assembly and the sample tray. The procedure is required after disassembling the system or when you exchange the sample handler, the sample hotel, the needle parkstation, the needle assembly, or one of the main boards.

This function is implemented in the drawer detection and in the needle exchange routine.


Steps to follow

There are various ways to perform the auto referencing procedure of the G7167A/B multisampler (via Lab Advisor, CDS, or the Instant Pilot).

All the methods first require the following preparation steps:

  1. Empty the workspace of the Multisampler
  2. Close all drawers properly
  3. Install two sample trays in each drawer without sample containers
  4. All drawers should have been properly configured
  5. The needle assembly is installed in the needle parkstation

Auto Referencing via Lab Advisor

  1. Connect the instrument to Lab Advisor
  2. Go to Instrument Control
  3. Go to the Controls of the Multisampler and find Configuration
  4. Press the "Clear Data" button 
  5. Reboot the sampler with the on/off button
  6. Reconnect the Lab Advisor go back to Instrument Control and Configuration
  7. Click the Detect Drawers / Auto Referencing button and follow the procedure.

Auto Referencing via CDS

  1. Open the CDS of the instrument

  2. Right-click in the Active Area of the Multisampler. This action will open a drop-down menu. Select 'Modify' and select 'Drawer Configuration' (Figure 1).


    Figure 1. Drawer Configuration in OpenLab CDS


  3. Follow the steps given by the software for the Autoreferencing procedure

  4. When done, click the 'Back' button.

Auto Referencing via the Instant Pilot

  1. Connect the instrument with the Instant Pilot
  2. Press More and then press Diagnostics
  3. Find the tests under the Multisampler. Perform the steps for the Autoreferencing procedure given by Instant Pilot.

Tip
Learn how to effectively maintain, troubleshoot, and operate your Agilent LC System:
Agilent 1260 Infinity LC Systems with OpenLab CDS e-learning path available from Agilent education
  • Maintenance
  • 1260 Infinity II
  • Analytical LC
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help