Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • Gas Chromatography
    • GC/MS
    • LC/MS
    • Liquid Chromatography
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Liquid Chromatography Portal
  • Agilent Knowledge Portal
Liquid Chromatography Portal
LC Articles Troubleshooting Agilent Pump Errors due to Blockage of a Pump Head
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • +LC Maintenance
    • -LC Troubleshooting
      • Troubleshooting High Pressure Issues in Agilent LC Systems
      • Troubleshooting Peak Area Variation Between Injections
      • Forced Cold Start Procedure
      • Troubleshooting Remote/Cluster Partner Missing Error
      • How to Access Diagnostic Buffer File from LC Modules for Advanced Troubleshooting
      • How to Free up Licenses After License Check Message on Agilent Lab Advisor Software
      • "Could not open channel RD" error message in Lab Advisor
      • -LC Pumps Troubleshooting
        • Troubleshooting Unstable Retention Time Issues - Cleaning the Multichannel Gradient Valve (MCGV) on Agilent Quaternary Pump Modules
        • Troubleshooting MCGV Cross Port Leaks in Quaternary Pumps
        • How to Troubleshoot no flow or no pressure issues: Binary/Quaternary/Isocratic Pumps - G131XX, and G711XX
        • No Pressure, No Flow, RT shift Issues: Cleaning Acetonitrile (ACN) Aggregations in Agilent Pumps
        • Troubleshooting Agilent Pump Errors due to Blockage of a Pump Head
        • How to Release a Stuck Inlet Valve in a G7120A High Speed Pump
        • Troubleshooting "Vacuum Limit Not Reached" Error in Agilent LC Pumps
        • Turning Off the Built-in Degasser Within a Pump Module
      • +LC Injectors Troubleshooting
      • +LC Column Compartments Troubleshooting
      • +LC Detectors Troubleshooting
    • +LC Best Practices
    • +LC Firmware

Troubleshooting Agilent Pump Errors due to Blockage of a Pump Head

Created by Max Schütte Max Schütte over 2 years ago | Last modified by Carlos Vargas Carlos Vargas 11 months ago

This Information Applies To: Agilent G1312X and G7112X binary pumps.


Issue:

A blockage before the pressure sensor, e.g. in a pump head, will not give a high pressure or overpressure system error. Instead, the pump may show one of various error messages due to the pump drive experiencing too much resistance during the piston movement:

  • "motor drive power" (error ID 2041, 2042)
  • "servo restart failed" (error ID 2201, 2211)
  • "stroke length" (error ID 2206, 2216)
  • "initialization failed" (error ID 2207, 2217)
  • "home position out of limit" (error ID 2203)
  • "home position not found" (error ID 2204). 

If your LC system shows any of these errors, then please follow these steps to try to locate and rectify the issue.


Steps to follow:

After each step, turn on the pump, start a normal pump flow, and then check if one of the errors listed earlier re-occurs or not.

  1. Remove the capillary between outlet valve and piston 2 (shown in purple in Figure 1).
  2. Remove the y-capillary going from the pump head outlet to the mixing chamber (shown in blue in Figure 1). 
  3. Remove the Outlet Ball Valve (OBV, shown in green in Figure 1).
  4. Loosen the Active Inlet Valve (AIV, shown in yellow in Figure 1).


Figure 1.
 Different parts to be checked for a blockage:
Capillary between OBV and Piston 2 (purple), Y-Capillary from outlet to mixing chamber (blue), OBV (green), AIV, and its cartridge (yellow)

If the OBV or the AIV are causing the error, you may try to clean them. Sonicate the OBV or the AIV cartridge (see Figure 1) first in isopropanol, then water and then in isopropanol, each for 10 minutes (do not sonicate the whole AIV, only the cartridge).

 Caution 
Caution: Remove the AIV cartridge from the AIV and only sonicate the AIV cartridge. Do not sonicate the AIV body.

If this procedure does not solve the issue, you may proceed with checking the pump head interior:

  1. To check if a piston is broken or if a piston seal is somehow twisted, remove the pump head and open the pump head.
  2. Check the pistons for deposited material that may be hindering its movement.
  3. Check if the spring in the pump head piston housing is broken (Figure 2).


    Figure 2. Piston Housing of the Pump Head Assembly: The red arrows show the springs pushing back the pistons. In the right part of the image, the glass piston is visible.

  4. Reinstall the pump head.

If the procedures described do not solve your issue or enable you to localize a blockage in your flow path, contact your local Agilent support representative for assistance.

 Tip 
Learn how to effectively maintain your Agilent HPLC:
Agilent 1260 Infinity LC Systems with OpenLab CDS Learning Path available from Agilent Education

  • Troubleshooting
  • Pumps
  • Analytical LC
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help