Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Liquid Chromatography Portal
  • Agilent Knowledge Portal
Liquid Chromatography Portal
LC Articles Troubleshooting Peak Area Variation Between Injections
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • +LC Maintenance
    • -LC Troubleshooting
      • Troubleshooting High Pressure Issues in Agilent LC Systems
      • Troubleshooting Peak Area Variation Between Injections
      • Forced Cold Start Procedure
      • Troubleshooting Remote/Cluster Partner Missing Error
      • How to Access Diagnostic Buffer File from LC Modules for Advanced Troubleshooting
      • How to Free up Licenses After License Check Message on Agilent Lab Advisor Software
      • "Could not open channel RD" error message in Lab Advisor
      • +LC Pumps Troubleshooting
      • +LC Injectors Troubleshooting
      • +LC Column Compartments Troubleshooting
      • +LC Detectors Troubleshooting
    • +LC Best Practices
    • +LC Firmware

You are currently reviewing an older revision of this page.

  • History View current version

Troubleshooting peak area variation between injections

This Information Applies To: All Agilent HPLC Autosampler modules


Issue:

Nonreproducible peak areas between injections, using the same sample and the same experimental conditions.


Resolution: 

The issue may come from different sources, it is recommended to verify all the following ones for a successful troubleshooting.

 Warning 
Warning: When opening capillary or tube fittings, solvents may leak out. Observe appropriate safety procedures (for example, wear goggles, safety gloves and protective clothing) as described in the material handling and safety data sheet supplied by the solvent vendor, especially when toxic, or hazardous solvents are used.
  • Vials
    Use only Agilent vials, the shape, and the material is optimized for maximizing instrument precision.
    Do not overfill the vials. Fill the vial up to 1/2 to 2/3 of the total volume.

  • Experimental conditions
    Make sure that the same solvents were used between the runs.
    Keep the room temperature constant and under 25 degrees Celsius.

  • Acquisition method settings
    Optimize the draw speed decreasing it down to 50µL/min. The default settings are at 200µL/min, but this speed may be too high for viscous samples.

  • Hardware
    The issue may reside in the Autosampler. A leakage is one of the frequent causes, make sure that all the autosampler connections are tight and there is no leaks between the needle and needle seat.

If the problem persists, contact Agilent Technologies for further guidance.