LC Articles Troubleshooting Peak Area Variation Between Injections
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • +LC Maintenance
    • -LC Troubleshooting
      • Troubleshooting High Pressure Issues in Agilent LC Systems
      • Troubleshooting Peak Area Variation Between Injections
      • Forced Cold Start Procedure
      • Troubleshooting Remote/Cluster Partner Missing Error
      • How to Access Diagnostic Buffer File from LC Modules for Advanced Troubleshooting
      • How to Free up Licenses After License Check Message on Agilent Lab Advisor Software
      • "Could not open channel RD" error message in Lab Advisor
      • +LC Pumps Troubleshooting
      • +LC Injectors Troubleshooting
      • +LC Column Compartments Troubleshooting
      • +LC Detectors Troubleshooting
    • +LC Best Practices
    • +LC Firmware

Troubleshooting Peak Area Variation Between Injections

Created by Paola Labarbuta Paola Labarbuta over 2 years ago | Last modified by Paubudge Paubudge over 1 year ago

This Information Applies To: All Agilent HPLC Autosampler modules


Issue:

Nonreproducible peak areas between injections, using the same sample and the same experimental conditions.


Resolution: 

The issue may come from different sources, it is recommended to verify all the following ones for a successful troubleshooting.

 Warning 
Warning: When opening capillary or tube fittings, solvents may leak out. Observe appropriate safety procedures (for example, wear goggles, safety gloves and protective clothing) as described in the material handling and safety data sheet supplied by the solvent vendor, especially when toxic, or hazardous solvents are used.
  • Vials
    Use only Agilent vials, the shape, and the material is optimized for maximizing instrument precision.
    Do not overfill the vials. Fill the vial up to 1/2 to 2/3 of the total volume.

  • Experimental conditions
    Make sure that the same solvents were used between the runs.
    Keep the room temperature constant and under 25 degrees Celsius.

  • Acquisition method settings
    Optimize the draw speed decreasing it down to 50µL/min. The default settings are at 200µL/min, but this speed may be too high for viscous samples.

  • Hardware
    The issue may reside in the Autosampler. A leakage is one of the frequent causes, make sure that all the autosampler connections are tight and there is no leaks between the needle and needle seat.

If the problem persists, contact Agilent Technologies for further guidance.

 Tip 
Learn how to effectively troubleshoot your LC autosampler:
HPLC-0GEN-2201e - Chromatographic Troubleshooting for HPLC e-learning course available from Agilent Education

  • Troubleshooting
  • injectors
  • Standard Autosampler
  • Analytical LC
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help