Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • Gas Chromatography
    • GC/MS
    • LC/MS
    • Liquid Chromatography
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Liquid Chromatography Portal
  • Agilent Knowledge Portal
Liquid Chromatography Portal
LC Articles General Column Regeneration Procedure
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • +LC Maintenance
    • +LC Troubleshooting
    • -LC Best Practices
      • Prevent Blocking of Solvent Filters
      • Good practices for use of the Multi-Channel Gradient Valve in Agilent quaternary pump modules
      • Mobile Phase Preparation for Isocratic LC Systems
      • General Column Regeneration Procedure
      • General Recommendations for Using an Agilent RID
      • How to Prepare your HPLC for a Long Time Shut Down
    • +LC Firmware

General Column Regeneration Procedure

Created by Shanice Rakotonindraina Shanice Rakotonindraina over 2 years ago | Last modified by Carlos Vargas Carlos Vargas over 1 year ago

This Information Applies To: Agilent LC Systems


Issue

After analyzing dirty or concentrated samples, some sample compounds may retain in the column and other parts of the HPLC system. If the column re-equilibration after each run is not enough to elute the retained compounds, you will notice several performance problems. 

Tailing or wider peaks, higher background signal, carry over, or ghost peaks can be a sign of a contaminated column. The use of low-quality solvents can also be the cause of those symptoms. 

As you use your column and expose it to various mobile phases, analytes, and sample matrices, these interactions will subtly affect it. Over time, there may be changes in its resolution. It’s important to keep a test chromatogram, from when your column was new, to compare and understand these changes in performance over time. If resolution degrades beyond an acceptable value for good quantitation, the column should be discarded and replaced with a new one.

Sometimes you can alleviate a high pressure or peak tailing issue by cleaning your column while back flushing it. Before cleaning your column, we recommend that you disconnect it from your detector and run your wash solvents into a beaker or other container. 

Check with your column manufacturer or review the instructions that came with your column before applying the following procedure. Start with the recommended cleaning or regeneration procedure from your column manufacturer. If it is not successful, you can use the procedure in this article as an alternative.


Steps to follow

 Caution 
Caution: Always use chromatographic grade solvents and wear the proper Personal Protection Equipment.

 Caution 
Caution: Methylene chloride, tetrahydrofuran, or dimethyl sulfoxide can swell PEEK components of the LC flow path. Confirm the material of the capillaries, fittings, or rotor seals before performing the following procedures. If any of the flow path components is made of PEEK, replace it or bypass it.

 Caution 
Caution: Check with your column supplier that one of the following regeneration procedures is suitable for your column and will not cause column damage before performing the following steps.

Regeneration Procedure for Reversed-Phase Silica Columns

  1. Disconnect the column and reconnect it to the system, with the flow through the column in the reversed direction. Do not connect the column to the detector to avoid contamination or clogging.
  2. Flush out any salts/buffers with HPLC grade water. Pump 25 mL of water through the column at 1 mL/min. For the following flushing solvents, use 1ml/min also.
  3. Flush column with 25 mL of isopropanol.
  4. Flush column with 25 mL of methylene chloride.
  5. Flush column with 25 mL of hexane.
  6. Flush column again with 25 mL of methylene chloride.
  7. Flush column again with 25 mL of isopropanol.
  8. Reconnect the column to the system, with the flow in the usual direction. Flush the column with the mobile phase without the buffer, then re-introduce the buffer.
  9. Equilibrate the column with 25 to 50 mL of mobile phase.
  10. To see if performance is restored, inject a standard or a sample.

Note: For some retained compounds that have fouled the column, dimethylformamide may be a better "cleaning" solvent than methylene chloride and hexane.

Regeneration Procedure for Normal-Phase Silica Column

  1. Disconnect the column and reconnect it to the system, with the flow through the column in the reversed direction. Do not connect the column to the detector to avoid contamination or clogging.
  2. Set the flow at 1 mL/min. Flush column with 50 mL of 50:50 methanol:chloroform.
  3. Flush column with 50 mL of ethyl acetate.
  4. Reconnect the column in the proper flow direction.
  5. Equilibrate the column with 50 mL of mobile phase.
  6. To see if performance is restored, inject a standard or a sample.
 Tip 
Tip: Follow the evolution of the signal and the background noise during the flush-out, especially before and after the regeneration procedure.

 Tip 
Learn how to effectively troubleshoot the Agilent LC Column:
HPLC-0GEN-2201e - Chromatographic Troubleshooting for HPLC e-learning course available from Agilent Education

  • Best practices
  • Analytical LC
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help