Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • Gas Chromatography
    • GC/MS
    • LC/MS
    • Liquid Chromatography
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Liquid Chromatography Portal
  • Agilent Knowledge Portal
Liquid Chromatography Portal
LC Articles How to Replace the PTFE Frit of the Purge Valve in LC Pumps
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • -LC Maintenance
      • -LC Pumps Maintenance
        • How to Replace the PTFE Frit of the Purge Valve in LC Pumps
        • How to Replace an Inlet Valve in a G7120A Agilent 1290 Infinity II High Speed Pump
        • How to change the Active Inlet Valve (AIV) and cartridge on an Agilent 1260 Isocratic, Binary or Quaternary Pump
        • How to Change the Passive Inlet Valve (PIV) on a 1260 Isocratic or Quaternary Pump
        • How to Change the Outlet Valve on a 1260 Isocratic or Quaternary Pump
        • How to Remove the Pump Head From an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump
        • How to change seals in a 1260 Infinity Binary, Quaternary, or Isocratic Pump without Seal Wash Option
        • Reinstalling a pump head on Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules
        • Seal Wear in Procedure for 1260 Isocratic and Quaternary Pumps
        • How to Sonicate the Inlet and Outlet Valves of Agilent 1100, 1200 and 1260 Series Pump Heads
      • +LC Injectors Maintenance
      • +LC Detectors Maintenance
    • +LC Troubleshooting
    • +LC Best Practices
    • +LC Firmware

How to Replace the PTFE Frit of the Purge Valve in LC Pumps

Created by Paola Labarbuta Paola Labarbuta over 2 years ago | Last modified by Xu Li-hua Xu Li-hua over 1 year ago

This Information Applies To: 1100, 1200 and 1260 Infinity I and 1260 Infinity II LC Pumps (Excluding preparative pumps with electromagnetic purge valve (EMPV): G1361A. Not applicable to any of the 1290 Infinity II modules with automatic purge valve: G4220X, G7104X, G7120X).


Issue:

Replacement of the Polytetrafluoroethylene (PFTE) frit is recommended every three months to avoid sample contamination and pressure issues.
Blockage of PTFE frits leads to increased system pressure. If high pressure is detected after opening the purge valve (around 10 bar), the PTFE frit is blocked and it must be replaced.


General overview:

Purge Valves (PV) are manual valves used to divert the flow from the column to the waste. A PTFE frit inside the purge valve catches solvent particles or seal wear.

Figure 1 illustrates the PV components.

 
Figure 1. Purge Valve Assembly components.
1. Seal cap, 2. PTFE frit, 3. Nut, 4. Purge valve body.

Steps to follow:

 Warning 
Warning: When opening capillary or tube fittings, solvents may leak out. Wear PPE as described in the material handling and safety data sheet supplied by the solvent vendor, especially when toxic, or hazardous solvents are used.
  1. Release the pressure by opening the Purge Valve, disconnect the outlet capillary and the waste tube.

  2. Remove the body of the Purge Valve (PV) from the Pump Head using a thin 14 mm wrench (p/n 8710–1924) to loosen the nut counterclockwise. Do not turn the body of the PV to unscrew it, since it is fixed on the pump head.

  3. Gently remove the Seal Cap with your hands without damaging it.

  4. Remove the PFTE frit using a needle, fine tweezers, or a flat screwdriver if necessary.

  5. Insert a new PFTE frit inside the PV. Make sure that the orientation is correct: the cut-side must face outside.

  6. Place the Seal Cap back by pressing it with your finger. If the Seal Cap is visibly damaged or deformed, then replace it with a new one to avoid leak issues.

  7. Assemble the PV body back into the Pump Head using a 14 mm wrench. Mind the orientation of the connection ports, it varies depending on the pump model (Figure 2). Hold the purge valve in the correct orientation to ensure the purge valve stays in this correct orientation during nut tightening.

  8. Reconnect outlet capillary and waste tube.

  9. Test the LC system backpressure with the purge valve open. The pressure value should now be less than the pressure value before replacing the filter.


    Figure 2. Correct orientation of PV connection port.

1. Legacy Pumps, 2. Agilent Infinity II LC Pumps.

 Tip 
Learn how to effectively maintain your Agilent LC Pump:
Best Practices for Efficient Liquid Chromatography primer on Agilent.com

 

  • Pumps
  • Maintenance
  • Analytical LC
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help