LC Articles How to Remove the Pump Head From an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • -LC Maintenance
      • -LC Pumps Maintenance
        • How to Replace the PTFE Frit of the Purge Valve in LC Pumps
        • How to Replace an Inlet Valve in a G7120A Agilent 1290 Infinity II High Speed Pump
        • How to Change the Passive Inlet Valve (PIV) on a 1260 Isocratic or Quaternary Pump
        • How to Change the Outlet Valve on a 1260 Isocratic or Quaternary Pump
        • How to Remove the Pump Head From an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump
        • How to change seals in a 1260 Infinity Binary, Quaternary, or Isocratic Pump without Seal Wash Option
        • Reinstalling a pump head on Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules
        • Seal Wear in Procedure for 1260 Isocratic and Quaternary Pumps
        • How to Sonicate the Inlet and Outlet Valves of Agilent 1100, 1200 and 1260 Series Pump Heads
      • +LC Injectors Maintenance
      • +LC Detectors Maintenance
    • +LC Troubleshooting
    • +LC Best Practices
    • +LC Firmware

How to Remove the Pump Head From an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump

Created by Damian O'Neill Damian O'Neill over 1 year ago | Last modified by paubudge paubudge 8 months ago

This Information Applies To: Agilent 1260 Infinity Binary, Quaternary or Isocratic Pumps.


Issue:

Removal of the pump head from the pump metering drive for maintenance on an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module.


Steps to follow

 Caution 
Caution: When loosening the pump head blots, make sure that the bolts are loosening evenly left and right. Do not loosen one pump bolt fully while leaving the other bolt tight as it may cause misalignment of the pump head and snapped pistons

1. Remove the tubing from your purge valve to your autosampler. (See Figure 1)

 
Figure 1. Remove the purge valve waste line and the capillary from the purge valve outlet.

2. Power off the pump and remove the waste tubing from the purge valve.

3. If you have an Agilent 1260 Infinity Isocratic or Quaternary Pump, remove the tubing to your passive inlet valve (PIV). If you have an Agilent 1260 Infinity Binary Pump, remove the tubing to your Active Inlet Valve (AIV).

4. Remove the tubing from your outlet ball valve on the top of the pump head to your damper. (See Figure 2)

5. Next remove the tubing from your damper to piston 2, which is located on the bottom of the pump head. (See Figure 2)

 
Figure 2. Remove tubing from the outlet valve and from damper to piston 2.

6. To separate the pump head from the metering drive, use the provided Allen wrench to loosen the pump head bolts 2 to 3 turns each, moving side to side, to make sure you are loosening the bolts evenly (see Figure 3). Only loosening one side can twist the pump head on the metering drive and can lead to piston breakage.

Caution 
Caution: In the next step, make sure to keep the rear of the pump head titled slightly up to prevent the piston failing out of the back of the pump head. 

7. Once the bolts are removed, slide the pump head gently off the metering drive. Be careful to keep the back of the pump head tilted upwards to prevent the pistons falling out of the pump head and breaking. (See Figure 3)

 
Figure 3. Removing the pump head from the pump metering drive.

Caution 
Caution: Do not remove the small bead of grease you see on the face of each piston at the back of the pump head. This is required for lubrication of the piston face against the metering drive.

 Tip 

Learn more on how to effectively troubleshoot your Agilent 1260 Infinity Pump:
HPLC-1260-2230z - Changing Pump Seals in Agilent 1260 Binary and Quaternary and Isocratic Pumps e-learning course available from Agilent Education

  • Pumps
  • Maintenance
  • 1260 Infinity
  • 1260 Infinity II
  • 1200 Series
  • Analytical LC
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help