LC Articles How to change seals in a 1260 Infinity Binary, Quaternary, or Isocratic Pump without Seal Wash Option
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • -LC Maintenance
      • -LC Pumps Maintenance
        • How to Replace the PTFE Frit of the Purge Valve in LC Pumps
        • How to Replace an Inlet Valve in a G7120A Agilent 1290 Infinity II High Speed Pump
        • How to Change the Passive Inlet Valve (PIV) on a 1260 Isocratic or Quaternary Pump
        • How to Change the Outlet Valve on a 1260 Isocratic or Quaternary Pump
        • How to Remove the Pump Head From an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump
        • How to change seals in a 1260 Infinity Binary, Quaternary, or Isocratic Pump without Seal Wash Option
        • Reinstalling a pump head on Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules
        • Seal Wear in Procedure for 1260 Isocratic and Quaternary Pumps
        • How to Sonicate the Inlet and Outlet Valves of Agilent 1100, 1200 and 1260 Series Pump Heads
      • +LC Injectors Maintenance
      • +LC Detectors Maintenance
    • +LC Troubleshooting
    • +LC Best Practices
    • +LC Firmware

How to change seals in a 1260 Infinity Binary, Quaternary, or Isocratic Pump without Seal Wash Option

Created by Damian O'Neill Damian O'Neill over 1 year ago | Last modified by Validated_paubudge Validated_paubudge 9 months ago

This Information Applies To: All Agilent 1260 Infinity Binary, Quaternary, or Isocratic Pumps without the Seal Wash Option installed.


Issue:

Worn pump seals can shed small particles that can cause blockages and pressure issues downstream of the seals. Agilent recommends that pump seals are changed yearly, or if pump performance indicates seal wear. Indications of seal wear include leaks at the lower pump head side, unstable retention times, or unstable pressure ripple.

Replacement of the pump seals in the 1260 pump head may be required in the following cases:

  • Due to a leak or some other pressure problem
  • A fail of the pressure test in Lab Advisor
  • Pump leak test in Lab Advisor suggests a problem with the pump seals
 Tip  Always call or check with your local service representative to confirm that seal replacement is required

Steps to follow

To change the seals you have to remove the pump head, see the following article for detailed instructions: How to remove the pump head from an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module 

 Warning 
Warning: Make sure that appropriate PPE is used for the chemicals used in this process. Nitrile gloves (or equivalent) and safety glasses are recommended.

 Caution 
Caution: Make sure that the piston housing is tilted upward when removing it from the pump housing, to prevent the pistons falling out potentially causing piston damage.


1. Unlock the lock screw and separate the piston housing from the pump housing. Keep the piston housing always tilted upwards to prevent the pistons from falling out of the piston housing. (See Figure 1)


Figure 1. Separate the piston housing from the pump housing. Take care to avoid pistons falling out of the piston housing.

2. Remove the support rings and seal holders to reveal the piston seal underneath.

3. Gently remove the pistons from the piston housing.

 Caution 
Caution: Do not remove the grease from the piston face and do not place the piston face down (Grease-side down) on the bench. The grease is to lubricate the face of the piston that is in contact with the pump drive, so must be kept in place to avoid wear and damage to both.

4. Clean each piston using isopropyl alcohol and lab wipes, then examine each piston shaft for damage or scratches (See Figure 2). If the pistons seem scratched or damaged, please immediately log a service request with your local Agilent service representative.


Figure 2. Cleaning the pistons

5. Using the steel end of the insertion tool, remove the pump seals. (See Figure 3)

6. Clean the inside the pump housing, the support rings, and the seal holders with HPLC-grade Isopropyl Alcohol (IPA) and laboratory wipes. (See Figure 3)


Figure 3. Removing old seals and cleaning pump head internal parts

7. Insert new seals into the pump housing using the plastic side of the seal insertion tool. (See Figure 4)


Figure 4. Inserting new piston seals into the pump head

8. Clean inside the pump housing, the support rings, and the seal holders with HPLC-grade isopropyl alcohol (IPA) and laboratory wipes.

9. Reinstall the seal holders and then the support rings onto the pump housing.

10. Line up the hole on the piston housing and the peg in the pump housing and connect the piston housing and the pump housing together.

11. Gently push the pistons into the housing making sure to align the shaft of the piston through the hole in the middle of each support ring and the pump seal. Also take care not to remove the grease from the face of each piston when inserting the piston down into the piston housing. (See Figure 5)

12. Tighten the lock screw the hold the two halves of the pump head together. (See Figure 5)


Figure 5. Pump head reassembly

13. The assembled pump head can now be mounted onto the pump. See Reinstalling a pump head on Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules for instructions on installing the pump head onto the pump module.

 Tip  Learn more on how to effectively troubleshoot your 1260 Infinity Pump Head please see a link to the following Agilent Education course:
HPLC-1260-2230z - Changing Pump Seals in Agilent 1260 Binary and Quaternary and Isocratic Pumps
  • Pumps
  • Maintenance
  • 1260 Infinity
  • Analytical LC
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help