Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Liquid Chromatography Portal
  • Agilent Knowledge Portal
Liquid Chromatography Portal
LC Articles Troubleshooting Remote/Cluster Partner Missing Error
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • +LC Maintenance
    • -LC Troubleshooting
      • Troubleshooting High Pressure Issues in Agilent LC Systems
      • Troubleshooting Peak Area Variation Between Injections
      • Forced Cold Start Procedure
      • Troubleshooting Remote/Cluster Partner Missing Error
      • How to Access Diagnostic Buffer File from LC Modules for Advanced Troubleshooting
      • How to Free up Licenses After License Check Message on Agilent Lab Advisor Software
      • "Could not open channel RD" error message in Lab Advisor
      • +LC Pumps Troubleshooting
      • +LC Injectors Troubleshooting
      • +LC Column Compartments Troubleshooting
      • +LC Detectors Troubleshooting
    • +LC Best Practices
    • +LC Firmware

Troubleshooting Remote/Cluster Partner Missing Error

Created by Max Schütte Max Schütte over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 7 months ago

This Information Applies To: All Agilent LC systems.


Issue:

A cluster is partnership between modules. Most modules are linked to other modules automatically, without intervention from the user, but sometimes the cluster is configured manually.

Examples:
1. To work as a ‘one module’ unit, independent modules must be clustered (e.g.: Two G1361A prep pumps are clustered to make one gradient pump).
2. A valve pressure cluster between the 1290 Infinity pump and external valves allows for correct operating pressure.
3. A sampler being assigned a specific pump, especially in a two-pump system.

If the cluster is changed or the communication is lost, the resulting error may be Remote Partner Missing or Cluster Partner Missing and the device will switch to Not Ready status.


Steps to follow:

  1. First Check that all modules are powered on, and the CAN cables are connected.
  2. In general, forcing a cold start should resolve the remote partner missing error. Check this article on how to do it: Forced Cold Start Procedure. 
  3. For autosamplers: A sampler may need the pump re-assigned (e.g. after an exchange of the pump), which may be done within the instrument software in the Control menu. Right-click the module icon and select Control. From the drop-down menu under Pump connected to sampler, choose which pump is connected.
  4. Firmware Declustering is available in the Lab Advisor Software as an option in Service & Diagnostics (please see Figure 1). A cluster can be declustered to avoid a Not Ready condition and a Cluster partner missing error. 

 Tip  Note: Firmware declustering was added to the Basic version of Agilent Lab Advisor since revision B.02.12 so if your version doesn't show it, update it. Copy-paste the following link in your browser to download the latest available version of Lab Advisor: http://labadvisor.pl29.agilent.com/latest 

 



Figure 1. Lab Advisor Firmware Declustering

During autoconfiguration, if the system detects that there are modules that can be clustered, a dialogue allows you to select the modules to include in the cluster via check boxes. Select the modules for the Column Compartment Cluster and the Pump/Valve Cluster separately.

To start the clustering, pick one of the available cluster option buttons (see Figure 2).



Figure 2: Autoconfiguration - Cluster Configuration

If this procedure does not solve your issue, please contact your local Agilent support representative.

 Tip 
Learn how to effectively operate your Agilent HPLC:
Agilent 1260 Infinity LC Systems with OpenLab CDS e-learning path available from Agilent education

  • Troubleshooting
  • Analytical LC
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help