This Information Applies To: Agilent 1260 Infinity II multisampler, 1290 Infinity II multisampler, 1260 Infinity II SFC multisampler, and 1260 Infinity II bio-Inert multisampler
Issue
The needle bends while moving to the wash port or showing one the following errors in the acquisition software:
- EE 25034,0: 'Sampler alignment procedure command failed'
- EE 25098,0: 'Alignment procedure: hotel command failed'
- EE 25151,x: 'Axis did not reach its target position in time'
- EE 25121,0: 'Sampler transport initialization failed'
Resolution
Try one of the following
- Make sure that the needle assembly is properly installed. If it is damaged, replace it.
- Run an autoreference. More details on the article How to Autoreference the G7167A/B Infinity II Multisampler
- Check the firmware version and update to the latest version compatible with your acquisition software. See the article How to Update or Downgrade Firmware for Agilent LC and CE Modules .
If none of the previous suggestions solve the issues, or the Sample Handler has been proven to be defective, please reach an Agilent Support representative.
|
Learn how to effectively troubleshoot your Agilent system:
HPLC-0GEN-2201e - Chromatographic Troubleshooting for HPLC e-learning course available from Agilent education
|