Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Liquid Chromatography Portal
  • Agilent Knowledge Portal
Liquid Chromatography Portal
LC Articles New UV Lamp Not Working - Lamp Tag Not Ready Error
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • +LC Maintenance
    • -LC Troubleshooting
      • Troubleshooting High Pressure Issues in Agilent LC Systems
      • Troubleshooting Peak Area Variation Between Injections
      • Forced Cold Start Procedure
      • Troubleshooting Remote/Cluster Partner Missing Error
      • How to Access Diagnostic Buffer File from LC Modules for Advanced Troubleshooting
      • How to Free up Licenses After License Check Message on Agilent Lab Advisor Software
      • "Could not open channel RD" error message in Lab Advisor
      • +LC Pumps Troubleshooting
      • +LC Injectors Troubleshooting
      • +LC Column Compartments Troubleshooting
      • -LC Detectors Troubleshooting
        • New UV Lamp Not Working - Lamp Tag Not Ready Error
        • Troubleshooting a failed ASTM Drift and Noise Test for the Agilent RID
        • How to Flush an HPLC Flow Cell to Troubleshoot Baseline Noise
        • Removing a Clog from a Flow Cell
    • +LC Best Practices
    • +LC Firmware

New UV Lamp Not Working - Lamp Tag Not Ready Error

Created by Max Schütte Max Schütte 8 months ago | Last modified by Carlos Vargas Carlos Vargas 2 months ago

This Information Applies To: Agilent Variable Wavelength Detectors G1314D, G1314E, G1314F, G7114A/B, and Diode Array Detectors G1315C, G1315D, G7115A, G7117A/B/C.


Issue:

When using UV lamps that do not have the RFID tag built in, the VWD and DAD modules may deliver an error upon exchanging the lamp. Errors can include "UV lamp tag not ready", together with the normal status "UV lamp not ready". This article describes how to resolve this issue.


Steps to follow:

  1. Go to instrument status view in your CDS software OpenLab ChemStation or OpenLab 2. The instrument status windows can be displayed by selecting Method and Run Control > Instrument Control or Status > Instrument Status > Dashboard, respectively. Right-click your detector module and select Control (Figure 1):



    Figure 1. Drop-down menu after right-clicking the VWD module

  2. In the control window, depending on your software, driver version, and module firmware, select the corresponding option for the UV Lamp Tag (Figure 2, Figure 3).

    The correct option depends on the lamp you inserted, but if installing a lamp without a RFID tag then the appropriate manual mode selection, or always use lamp option can stop the above errors from re-occurring.



    Figure 2. Example of the control window for a G1314F in OpenLab 2 - select manual mode for lamps without tag



    Figure 3. Example of the control window for a G1365C or G1315D in OpenLab 2 - select always use lamp

If your new lamp still does not work and you have other compatible detectors, try this lamp in a module that is currently running error-free. If the lamp works, the problem lies with your detector. If the lamp does not work, the problem lies with the lamp. 

If this procedure does not solve your issue, please contact your local Agilent support representative.

 Tip 
Learn how to effectively maintain and operate your Agilent 1200 Infinity Series VWD:
https://www.agilent.com/cs/library/usermanuals/public/G1314-90034_VWD_DEF_USR_EN.pdf user manual on Agilent.com
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help