This Information Applies to: Analytical-scale LC systems with a single pump with manual purge valve. It does not apply to systems with Agilent G7167A/B Multisampler or extra switching valves.
Issue
A low pressure through the column is observed, however, no leak is detected or visible in the system.
Steps to Follow
General system considerations
Sampler
Visually check the sample flow path for leaks. Salt deposits may be an indicator of small leaks.
Tips: Touching those areas with a paper towel can be helpful. Using LabAdivsor Maintenance commands to have better access to the needle is recommended.
|
If one of the previously mentioned parts is leaking, try retightening first. If they are bad condition, replace them.
Pump
Caution: If a seal wash option is installed, turn it off before continuing with the troubleshooting steps.
|
Figure 1. Pump head.
1. Pump outlet capillary, 2. Purge valve, 3. Waste tube
Figure 2. Pump head.
1. Pump outlet capillary, 2. Purge valve, 3. Waste tube
If the previous procedures did not solve the issue, or you were not able to perform them, contact your local Agilent Service Representative for assistance.