This Information Applies To: Agilent 1290 Infinity and 1290 Infinity II LC Pumps.
Issue
Solvent bottle runs dry. As the result, the pump has continued to run dry or solvent lines are filled with air. The difference between purge, condition, and prime functions and when to use them are explained.
Caution: Running a pump dry can damage the pistons and wear out the piston seals prematurely. Avoid this issue by stopping the pump flow when out of solvent. See How to Stop Pump Flow When Out of Solvent for further details.
|
Items Required
Warning: This procedure risks exposure to hazardous, toxic, or flammable solvents and reagents. Wear appropriate personal protective equipment as described in the material handling and safety data sheet supplied by the chemical vendor, and always follow good laboratory practice.
|
Steps to Follow
Tips: With PPE on, place paper towel underneath the solvent selection valve (SSV) and tubing to the degasser to catch any drips and avoid setting off the module's leak sensor.
|
Caution: Never use the Prime function to fill empty solvent lines connected to the pump head. The process is stressful to the purge valve and seals. Use a syringe to manually fill empty solvent lines.
|
Figure 4. Priming degasser in Flexible/Quaternary pumps
Figure 5. Priming an SSV in High Speed/Binary pumps
Tips: See Best Practices for Using an Agilent LC System for more information.
|